Precise 111m tube tester restoration
#16

very nice troubleshooting! Great to be able to reason out the problem. I'm more of a trial and error experimentalist Icon_lol
#17

Thanks Bob! This has been a challenge. A little different than a radio, but hopefully this will play a hand in the restoration of lots of radios, TVs and other vintage electronic gear. I got to studying on the jumpy meter. It seems this meter is very sensitive, and likes to act like a windshield wiper when it starts to measure, bouncing back and forth several times before settling. It also seems a little unstable during a test. Bugs me. So, I set up a Simpson meter beside it and watched things. It looks like it is tracking the line voltage. Maybe even amplifying it. Someone from ARF mentioned his was the same way, so I take it this is the personality of this tester. I have been through it, checking for bad solder connections and dirty switch contacts. I also noticed my old 111 and Jackson did about the same thing.

 So, I did some experimenting and came up with this: I found that a large value, low voltage electrolytic capacitor put parallel to the meter (across the existing 100mf cap) made a world of difference in the behavior. I found that a 1000mf on up made things behave a lot better. I had the best results using a 4700mf, 10v cap. The best place I found to connect this was across the center lugs of the meter reverse switch.
   
It seems to be working really well.
I have calibrated it, and need to move on to the cabinet. I am still looking for any tube data updates.

If I could find the place called "Somewhere", I could find "Anything" Icon_confused

Tim

Jesus cried out and said, "Whoever believes in me , believes not in me but in him who sent me" John 12:44
#18

I have the cabinet finished, things put back together and have tested several tubes. It seems fairly accurate for the most part, no tubes unusually low or high, other than a 6BQ5 tube, and I think the settings or something must be wrong in the chart, since even new ones test about half the minimum. The old 111 reads the same. Other audio output tubes test ok. Other than that, it works well. So, unless I need to recalibrate because of an error. What kind of meter to use is still a bit of a mystery to me, I used a Simpson 260, and a Fluke where it called for a VTVM.

 So, I pronounce it finished. Icon_wink
   


A few notes:

I have the complete construction manual.
Have 'improved' the schematic and parts layout by adding the part location numbers (R1, C2, etc.)
If anyone needs a copy, PM me.

Added 4700 mf cap across the meter - flat tamed it down and helps eliminate slamming the meter.

Found the primary of the power transformer reversed causing a no meter reading on Em test on anything using the D" (50V) setting on screen voltage. Don't know if this is a factory defect in the power transformer or an error on the schematic and build sheet.

Be very careful when removing or inserting into case. The test switch is exposed and can be broken. I recommend putting the face of the tester down, then gently and carefully remove or install case.

If I could find the place called "Somewhere", I could find "Anything" Icon_confused

Tim

Jesus cried out and said, "Whoever believes in me , believes not in me but in him who sent me" John 12:44
#19

An outstanding job on a fine instrument. You will be pleased with it's performance!
#20

UPDATE:

I realized that I had found and made notes on a couple errors on tube settings and forgot to update my thread. Hope this helps someone.

50L6 settings are wrong. Use the 25L6 setting and 50v heater. This is on most charts.

6BQ5 settings are wrong, use 8BQ5 settings with 6v heater. This is not on all of the Precise tube charts, so here are those settings:
   

If I could find the place called "Somewhere", I could find "Anything" Icon_confused

Tim

Jesus cried out and said, "Whoever believes in me , believes not in me but in him who sent me" John 12:44




Users browsing this thread: 1 Guest(s)
[-]
Recent Posts
Philco 91 Speaker Replacement
The spider is glued on the cone and so is the voice coil. I think you can use a cotton swab dipped in acetone and carefu...RodB — 02:42 PM
Restoring Philco 37-604C
Yes I saw that. If I knew, I'd probably try to arrange for getting it myself. But I have just got one. Which does not m...morzh — 01:44 PM
Philco 38-2 Automatic Tuning
When you push the lever, you are supposed to rotate the disk to the desired station. Then the magnetic tuning will acqu...morzh — 01:42 PM
Philco 38-2 Automatic Tuning
I figure out the muting from another picture that helped. Now I need to know how the automatic tuning works. When I push...dconant — 12:11 PM
Philco 91 Speaker Replacement
How would I go about removing the spider from the cone?dconant — 12:04 PM
Philco 38-2 Automatic Tuning
Hello, I am having trouble getting my automatic tuning to stay unmuted. If I play with the tuning handle I can get it to...dconant — 10:51 AM
Does anyone make photofinish replacements?
Here’s the link to the DIY photofinish section of our site: klondike98 — 09:46 AM
Delco car radio Peko vibrator converstion issues
Hi Richard, Thx for posting the schematic.  As I remember as a kid, these were great performers for the time, likely...MrFixr55 — 08:56 AM
Delco car radio Peko vibrator converstion issues
hello mr Fixr, The Electro  Powersupply that I am powering the radio with has a huge choke plus a 10,000ufd electrolyti...radiorich — 10:58 PM
Delco car radio Peko vibrator converstion issues
How about putting a choke in the power supply? The big cap was likely an attempt to stabilize the DC. Kids who cre...MrFixr55 — 09:19 PM

[-]
Who's Online
There are currently no members online.

>