02-01-2016, 11:24 AM
(01-31-2016, 05:36 PM)Benny/W5TYZ Wrote: ........wondered about the statement on the end "One side common to line and load". Really what are they saying?......
A common variac does not provide isolation as this general diagram shows (neutral is common to the both line and load):
[Image: http://www.afcaforum.com/attachment.php?id=53827]
Unless you buy something like a PR57 or BK 1655 that has a built in isolation transformer you'll need to use an isolation transformer in-line with a variac for true (and safe) isolation.
I guess it bears repeating for others that when you buy an isolation transformer don't automatically assume it is isolated. Ohm it out with a meter (while unplugged of course) to make sure everything is isolated. Some recent medical grade isolation transformers are actually a leakage reduction transformer.
Like morzh I bought a Dale IT400 and while checking it out discovered like he did that it had a jumper that had to be removed to make it an isolation transformer. Morzh documents his experience here:
http://www.philcoradio.com/phorum/showth...p?tid=4698
John KK4ZLF
Lexington, KY
"illegitimis non carborundum"