Variac / Variable Transformer.
#1

I have been looking for a Variac with isolated output.  
I have yet to find one.  Staco has a model 3PN J401,
that is output isolated.  I found one for $289.00. 2 much.

What is involved in going inside a regular unit and set
it up as Isolated Output?
 
I found a couple that have a statement on the exterior
next to the output that states."One side common to line and load" 
What does this mean?
Would that be the Neutral line is common all the way to the
mains?  The hot side is the one being transformed...Right??

Benny/W5TYZ
Orange, Texas
The second mouse gets the cheese
#2

I believe the Sencore PR-57 models are single unit isolated variacs. Those tend to be pricey though because they are intended to test different modes of isolation failure as well.

There's no reason you can't use a separate isolation transformer after the variac (before would work too I think). Just plug the variac into the wall, the isolation transformer into the variac, and the radio into the isolation transformer.

When I was starting out I bought something like this:
http://www.ebay.com/itm/Variac-Variable-...1872830853
A couple things I learned since then: the meter isn't useful (too small). And 20A is really overkill. One with a 5A or so rating would be plenty for radio work. Also, don't trust the dial markings - I just use my DMM to monitor the voltage.

I made an isolation transformer box using one of these that I got off of ebay for ~$50:
http://www.digikey.com/product-detail/en...ND/1118196

There are also isolation transformer boxes made for hospitals and labs that show up on eBay too.
#3

Thanks for the info Nathan.  I still have an isolation Xfmr.
I sold my Variac several years ago and never replaced it.
I thought this time I would try to find one with isolated output.
I thought maybe someone may have reworked a Variac to
make it isolated output.  I may just pickup a 5 or 10 amp
Variac, and use the Isolation Xfmr with it.
I have looked at the PR-57's, and they ARE pricey....
I guess if this was a sho'nuff repair business, the investment
would pay for itself.  After all it is a hobby.  We have a local
Hamfest here the end of next month, and just maybe I'll find
something useful.

Benny/W5TYZ
Orange, Texas
The second mouse gets the cheese
#4

Benny, I have been using an RCA-Isotap II WP-27A. It isn't continuously variable but instead you can change output voltage in 5 volt steps starting at 25 volts all the way up,to 150 volts and it is an isolation transformer too! I think a different version on of this is a VIZ-IsotapII WP-27A. I've seen them on eBay for around $40.

Ron

Bendix 0626.      RCA 8BX5.   RCA T64
Philco 41-250.    Philco49-500
GE 201.             Philco 39-25
Motorola 61X13. Philco 46-42        Crosley 52TQ
Philco 37-116.    Philco 70
AK 35                Philco 46-350
Philco 620B.       Zenith Transoceanic B-600
Philco 60B.         Majestic 50
Philco 52-944.    AK 84
#5

OK Ron...
Thanks for the tip.

Benny/W5TYZ
Orange, Texas
The second mouse gets the cheese
#6

Here is some good info on an isolation transformer from morzh. might give you some info your wondering bout.

http://philcoradio.com/phorum/showthread.php?tid=4698
#7

I gave up trying to find a PR57 because they were all going for more than I wanted to pay.  I did find a similar BK1655 on eBay a couple of years back that was much more reasonable, so you might want to search for one of those.  In the mean time pairing your isolation xfrmer with a variac is a good solution.

I have an old General Radio Company W5MT3 variac that is built like a tank that I bought at a hamfest several years ago pretty cheap.  I still see them on eBay but they aren't quite as cheap anymore:

http://www.ebay.com/itm/General-Radio-Tr...SwpdpVcQVj

John KK4ZLF
Lexington, KY
"illegitimis non carborundum"
#8

I have looked at the W5MT3 models and wondered about the statement
on the end "One side common to line and load". Really what are they
saying?

I am still working on radios, and looking for something reasonable that will
still do the job. Looks like finding a Variac to go with my isolationXfmr, will
be the best way to go for now......

Benny/W5TYZ
Orange, Texas
The second mouse gets the cheese
#9

(01-31-2016, 04:36 PM)Benny/W5TYZ Wrote:  ........wondered about the statement on the end "One side common to line and load".  Really what are they saying?......

A common variac does not provide isolation as this general diagram shows (neutral is common to the both line and load):

[Image: http://www.afcaforum.com/attachment.php?id=53827]

Unless you buy something like a PR57 or BK 1655 that has a built in isolation transformer you'll need to use an isolation transformer in-line with a variac for true (and safe) isolation.

I guess it bears repeating for others that when you buy an isolation transformer don't automatically assume it is isolated.  Ohm it out with a meter (while unplugged of course) to make sure everything is isolated.   Some recent medical grade isolation transformers are actually a leakage reduction transformer.

Like morzh I bought a Dale IT400 and while checking it out discovered like he did that it had a jumper that had to be removed to make it an isolation transformer.  Morzh documents his experience here:

http://www.philcoradio.com/phorum/showth...p?tid=4698

John KK4ZLF
Lexington, KY
"illegitimis non carborundum"
#10

I'm gonna throw a monkey wrench in this line of thought. I believe it is a good idea to have BOTH a separate variac and isolation transformer. Why? You NEED the protection of the isolation transformer when working on a bare chassis. You need to be able to vary the voltage to bring up, test and watch with the variac. In the real world, a radio/tv or such is designed to work directly off the line. My point? When you want to test something, there ARE times when you NEED to connect it directly to the line, but still have the variability of the variac. On the ac/dc radio sets, the chassis is connected directly to one side of the ac line. Fine when it is plugged in correctly. If it is reversed, then the chassis becomes "hot", and can give you a pretty good shock. The variac connects directly to the ac line, one side is simply a short from your outlet to the variac outlet, the other side runs through the transformer. It works like a dimmer switch that are on ceiling lights. So, you get NO protection using only a variac. Besides, if the variac or isolation trans fails, you only need to replace one or the other. I think it is good plan to have both. My experience anyway. Icon_e_biggrin

If I could find the place called "Somewhere", I could find "Anything" Icon_confused

Tim

Jesus cried out and said, "Whoever believes in me , believes not in me but in him who sent me" John 12:44
#11

Here is my retro-style homebrew iso-var. 

Will handle 1kW as a bare variac, 375VA when isolated up to 600V.

I use it for many tests, not just bugging out receivers.

Dennis



Attached Files Image(s)
           

Pacing the cage...
#12

I lucked out and found this Variac on Ebay, that has the isolated output.  I went to the company and downloaded the wiring diagram.  When it gets here, I'll check the circuit and make sure no one has changed anything.  I was the only bidder at $49.99, so that made it even better.


Attached Files Image(s)
   

Benny/W5TYZ
Orange, Texas
The second mouse gets the cheese
#13

Well, Suppose you have two big surplus 115 to 7.5 or whatever hefty surplus transformers and run 'tm back to back, you would get needed isolation, and then hookup your variac whether isolated or not, and do your testing. Yeah, well, I think the next stupid trick may be my last, so it takes me a little longer bench time to sort out the problem of the day. "Belt and suspenders, Gents, we want you here next year!"




Users browsing this thread: 1 Guest(s)
[-]
Recent Posts
Cannot Adjust an IF Transformer’s cores
It works! I received the PCBs this morning in the mail, and put everything together. I reused the base from the original...jrblasde — 04:32 PM
PT-6 antenna connections
I seem to have lost my way on this PT-6. I see the antenna aerial transformer schematic - I’m reading pins 1 through ...bridkarl — 12:30 PM
40-150 General questions
Good thing you're a good electrician ... the pictures show a lot of rewiring in your future! Take care and BE HEALTHY!...GarySP — 11:52 AM
40-150 General questions
I think RobB has it right. It is a part number and looks like its stamped on the chassis and that wasn't something you ...klondike98 — 11:41 AM
40-150 General questions
It's most likely a part number. Most of Philco part numbers are in that format. I looked at the parts list for this mode...RodB — 08:10 AM
40-150 General questions
Hi Kevin, No, I don't but I like your Quote about being a good electrician. The way I look at it is "The fact ...MrFixr55 — 08:06 AM
Philco 38-7 Speaker
7K should be fine. While not a perfect match (This transformer is meant for 6V6 tubes), Antique Electronic Supply (AES)...MrFixr55 — 08:03 AM
40-150 General questions
I discovered a stamp on the underside of the breadboard on my Philco 40-150. It is"56-1417' in black ink with numbe...Kevin_D — 07:14 AM
Philco 38-7 Speaker
FYI the Impedance of the HR20 (the one I have) secondary of the audio transformer is 5 ohms. The 38-7XX speaker is suppo...tludka — 04:22 PM
Philco 38-7 Speaker
Yeah, don't use the transformer currently on the H-20 (Dopf, I thought "H20, is that new technology, a speaker usin...MrFixr55 — 03:51 PM

[-]
Who's Online
There are currently 2231 online users. [Complete List]
» 1 Member(s) | 2230 Guest(s)
Avatar

>