Triplett 666 Resistors
#1

I think I found a schematic that may work for my 1939 666. The resistors are stamped and match the schematic but the meter operates differently. Its looks like on the 10v scale the 9900 ohm resistor is in place but to get the meter to operate properly, it needs a 2.3k resistor, not the 9900. I really dont want to re-resistor the whole meter. What could be the issue


Attached Files Image(s)
       
#2

Why 2.3K?

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#3

Thats the mystery question. I used a variable pot to calibrate it for a correct reading on the 10v scale and thats what I came up with
#4

Do you have the same exact galvanometer head? 400uA full deflection current? If it is the same, the head is out of order, if it is not, you need to know the parameters.

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#5

It appears to be the original meter but it does not say on the face plate. How can I apply 400 micro volts to see the full deflection?


Attached Files Image(s)
   
#6

400 micro-Amps, not micro-Volts.

Full deflection current will move the needle to the far right when put through the spindle coil, to which tthe needle is attached.
Then it is simple Ohms law: your full resistance (including the one of the meter itself, though it is rather small) will go into I=V/R formula, where the V is your voltage.
You could simply measure the resistance between the two contacts of your resistor (with the resistor itself disconnected) to see if it is 250 Ohms, or it is more than that. If it is more than that, you need to trace all the resistors that are in series with that resistor in question, and then your resistor will be

R-needed = (10V/0.0004A) - R-measured, or rather R-needed = 25kOhm - R measured.

If your R-measured is more than 250 Ohm, you need to find out what it is and why it is a part of your circuit, which (your colored part) shows straight wires.

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#7

I am lost. Where did you come up with 250 ohms?
#8

This is your schematic. It says under your galvanometer: 400uA, 250 Ohm.
It means that

1) 400uA will fully deflect your spindle.
2) 400uA will develop 0.1V dropout across that spindle coil.
3) this resistance is in series with any other resistance that is a part of your switched divider. Simply speaking, if there is nothing else in parallel or in series with this galvanometer, you will need 25kOhm resistor in series to fully deflect the needle at 10V.

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#9

ok, thanks, I will check it out.
#10

Looks like 1.8kohms @ 10v is full scale
#11

Have you measured the resistance of the circuit where that 9900 ohm resistor connects to (or where you connect your 1800 ohms one)? With your resistor disconnected?

And, the last thing: is there a shunt of some sort connected across the leads of the galvanometer?

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#12

I think the metering wiring is messed up. I disconnected the wires to the movement. No visable shunt. Measured from the probes to the meter connections. Same readings on both. Neg= continuity. Pos=12k. I dont understand it.
#13

Well, you have to trace it.
But that would explain why your resistor is much lower than what it should be according to the sch.

Does the soldering and all look factory-made, or are there signs of tampering?

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#14

https://bama.edebris.com/manuals/triplett/666h

https://elektrotanya.com/triplett_model-...ad.html#dl
(pass Captcha, wait a few sec and download)

These sch are a bit different

People who do not drink, do not smoke, do not eat red meat will one day feel really stupid lying there and dying from nothing.
#15

I dont have the R model, its much newer. Mine is the first one made in 1939, 666 but the H model looks similar




Users browsing this thread: 2 Guest(s)
[-]
Recent Posts
Philco 46-420 Code 121 Reception issues
Welcome Eric, I agree with Bob and far as the two main electrolytic filter capacitors did you change them yourself or w...radiorich — 11:43 PM
Philco 46-420 Code 121 Reception issues
You mentioned the Philco manual and going through the check points...just to be sure we're on the same page here's the m...klondike98 — 08:13 PM
Philco 42-1008 conversion kit
Interesting. I haven't seen that before.klondike98 — 07:02 PM
12' Philco
Yes I had looked for it on the web as well some time back and could not find it. I was glad to see it turned up in Ron'...klondike98 — 06:59 PM
Shadow Meter Bulb
Now if you had a set with a tuning light then the bulb type is important to the circuit, some sets used those prior to t...Arran — 04:58 PM
Shadow Meter Bulb
Ok. Thanks for the correction.RossH — 03:09 PM
Model 28L
For 28 you will probably need to buy a Hammond 125CSE. Or any of the series of the power you need, with SE suffix. Then ...morzh — 02:09 PM
37-60 revision 6
I am restoring a Philco 37-60 and it shows run 6 they removed the ground from G3 of the 6K7G and put the G3 to -2.5v for...bobbyd1200 — 01:01 PM
Shadow Meter Bulb
Mike is correct on the bulb connection, two separate circuits. I found that by rotating the bulb and sliding it forward ...RodB — 12:19 PM
Hickok AC51 tube tester
Cleaned ann contacts, switches and sockets, works great now.martinj — 11:32 AM

[-]
Who's Online
There are currently 6182 online users. [Complete List]
» 4 Member(s) | 6178 Guest(s)
AvatarAvatarAvatarAvatar

>